Our full technical support staff does not monitor this forum. If you need assistance from a member of our staff, please submit your question from the Ask a Question page.


Log in or register to post/reply in the forum.

Thermocouple reading sudden drop to reference temperate


TLT Apr 14, 2026 02:55 PM

Hi Everyone,

We have an issue on our new CR310 configured to data log 32 numbers of T-type thermocouple. The reading suddenly drop to a reading very close to reference temperature. 

Our CR310 was wired up as in 16/32B manual putting a T107 probe at the multiplexer. Please refer to 16/32B manual, " FIGURE C-2. Differential thermocouple measurement with reference junction at the AM16/32B (using 107-L thermistor)".

Previously, we had 2 sets with the same setup, same program were working fine, the only different this time we use another type of 4G industrial router.    

Sub ReadMux1
PortSet (1,1)
For Count1 = 1 To 32
PulsePort (2 ,10000)
TCDiff (Temp(Count1),1,mv34,1,TypeT,T107_C,True ,0,60,1,0)
Next
PortSet (1,0)
EndSub

TIMESTAMP RECORD batt_volt PTemp T107_C Temp(1) Temp(2) Temp(3) Temp(4) Temp(5) Temp(6)

11/4/2026 16:00 1183 12.36 38.38 34.44 81.2 32.57 32.78 33.23 77.37

11/4/2026 16:15 1184 12.43 38.21 34.25 81 32.22 31.93 32.47 77.22

11/4/2026 16:30 1185 12.44 37.99 34.04 80.9 31.59 31.14 31.48 77.05

11/4/2026 16:45 1186 12.4 37.75 33.83 80.8 31.79 31.46 31.78 76.94

11/4/2026 17:00 1187 12.25 37.58 33.65 80.7 31.55 31.11 31.41 76.8

11/4/2026 17:15 1188 12.44 37.42 33.49 80.5 31.39 30.94 31.27 76.69

11/4/2026 17:30 1189 12.33 37.22 33.3 80.4 31.27 30.64 30.85 76.55

11/4/2026 17:45 1190 12.32 37.01 33.1 80.3 30.79 30.14 30.45 76.43

11/4/2026 18:00 1191 12.33 36.74 32.84 63.92 30.19 28.18 28.67 69.58

11/4/2026 18:15 1192 12.33 36.22 32.44 32.43 32.44 32.44 32.44 32.43

11/4/2026 18:30 1193 12.42 35.45 31.85 59.6 28.8 27.12 27.49 71.02

11/4/2026 18:45 1194 12.34 34.86 31.35 72.73 28.61 27.17 27.51 75.85

11/4/2026 19:00 1195 12.25 34.52 30.95 30.94 30.95 30.95 30.96 30.95

11/4/2026 19:15 1196 12.41 34.13 30.61 33.79 27.36 30.38 28.33 31.8

11/4/2026 19:30 1197 12.37 33.75 30.19 30.2 30.2 30.19 30.19 30.19

11/4/2026 19:45 1198 12.39 33.43 29.89 29.9 29.9 29.89 29.89 29.89

11/4/2026 20:00 1199 12.24 33.24 29.66 29.65 29.65 29.65 29.65 29.65

11/4/2026 20:15 1200 12.36 33.11 29.48 29.48 29.49 29.48 29.48 29.48

11/4/2026 20:30 1201 12.34 32.9 29.29 29.29 29.29 29.29 29.3 29.29

11/4/2026 20:45 1202 12.24 32.59 29 29 28.99 28.99 29 29

11/4/2026 21:00 1203 12.29 32.33 28.73 28.72 28.72 28.73 28.73 28.72

11/4/2026 21:15 1204 12.29 32.22 28.58 28.58 28.58 28.58 28.58 28.56

11/4/2026 21:30 1205 12.29 32.2 28.49 28.49 28.49 28.49 28.49 28.48

We reversed the OS version to 11.01, due to the CR310 running at OS version 11.03 will send the data again after the CR310 reboot.

Anyone can help me on this. Please advice.

Thanks,

TLT


JDavis Apr 14, 2026 04:50 PM

For a reading to be near the reference, the voltage of the thermocouple is reading near 0 volts. Being flat across the readings, I think the multiplexer is not advancing channels. A loose connection could do that. Another possibility is not sending it correct signals.

Try with a delay of 150ms right after the first PulsePort. That will make sure the AM16/32B uses the old clocking mode instead of the new fast signalling mode.

Alternative is to use the MuxSelect instruction. The example programs online for the AM16/32B use that instruction.


TLT Apr 17, 2026 02:54 AM

Hi JDavis,

Thank you for your reply.

We checked the connection wire between CR310 and AM16/32B they were okay and tight. We used 22AWG signal wires. Is this wire size suitiable? 

What will be the possibility for not sending the correct signals could it be improper install of thermocouple wires?

We tried shorting the DIFF H and L terminals on the CR310 where the AM1632B COM ODD H and L connected, the readings became close to the reference temperture.

What are the possible cause the AM16/32B from not advancing channel?

Is this MUXselect instruction good enough to ensure that the AM16/32B will advance channel?

MuxSelect (C2,C1,20,1,1)
Count1=1
SubScan(0,uSec,32)
TCDiff (Temp(Count1),1,mv34,1,TypeK,T107_C,True,0,60,1,0)
Count1=Count1+1
PulsePort(C2,10000)
NextSubScan

PortSet(C1,0)

We are unsure about the frequency filtering selection. For now we are using 60Hz, however our local AC supply is 50Hz. We tested both frequency and we don't see a different. Is it becasue we are not using AC to DC converter to charge or power the CR310? Usually, we use a 12 volts battery and solar panel to do the charging.

   

Looking forward for you advice.

Regards,

TLT


JDavis Apr 17, 2026 10:27 PM

The code should work that way.

With measuring reference temperature at the multiplexer, do not use thermocouple wire. Keep using copper wire for both high and low of the differential voltage.

On a CR310, you won't see a difference between the 50 and 60 fNotch settings. They both do the same mode that rejects both 50Hz and 60Hz noise. On a CR1000Xe, which has a different analog converter, you would see a difference.


TLT Apr 22, 2026 03:45 AM

Hi JDavis,

Thank you for your reply.

We will keep a lookout when working with CR1000Xe.   

Does the noise come in only when we use AC to DC charging? How do we determine which filter to be use, do we follow the same Hz as our local AC power supply?

Looking forward for you advice.

Regards,

TLT


JDavis Apr 22, 2026 02:44 PM

Use the option that matches the frequency of local AC power when near any AC electrical wiring.The wires in the walls transmit noise even if the data logger is not plugged in to AC.

Even away from AC, the measurement is improved some using 50 or 60 fNotch versus a fast speed. The data logger specification sheet has a table for comparison.


TLT Apr 29, 2026 02:26 PM

Hi JDavis,

Thanks for your advice.

The frequency table is too complicated for me.

I will follow our local AC frequency 50Hz.

Thanks.

Regards,

TLT

Log in or register to post/reply in the forum.